- TL;DR Summary
- Source Specification
Can I input the X-Ray tube voltage in MCNP source specification?
One would put in energy or energy spectrum for a photon source. I am not aware that voltage for an X-ray tube is an input.Anisur Rahman said:TL;DR Summary: Source Specification
Can I input the X-Ray tube voltage in MCNP source specification?
Even if the current value (mA) differs?Alex A said:An electron falling through a potential difference of 1V gains 1eV of kinetic energy.
So an X-ray tube powered at 100kV would be simulated with a beam of electrons of energy 0.1 MeV (SDEF ERG= are in MeV) directed at a target.
The current determines the number of particles (mA/e) per sec, were e is the unit charge.mariem said:Even if the current value (mA) differs?